SETUP Menu Teledyne API - T100 UV Fluorescence SO2 Analyzer
110
5.9.1. SIGNAL I/O
The signal I/O diagnostic mode allows a user to review and change the digital and
analog input/output functions of the analyzer. Refer to Appendix A-4 for a complete list
of the parameters available for review under this menu.
IMPORTANT
IMPACT ON READINGS OR DATA
Any changes of signal I/O settings will remain in effect only until the
signal I/O menu is exited. Exceptions are the ozone generator override
and the flow sensor calibration, which remain as entered when exiting.
Access the signal I/O test mode from the DIAG Menu (refer to Figure 5-16), then press:
EXAMPLE
DIAG SIGNAL I / O
PREV NEXT JUMP ENTR EXIT
DIAG I / O 0) EXT_ZERO_CAL=OFF
PREV NEXT JUMP PRNT EXIT
Press JUMP to go
directly to a specific
signal
See Appendix A-4 for
a complete list of
available SIGNALS
DIAG I / O JUMP TO: 12
1 2 ENTR EXIT
EXAMPLE:
Enter 12 to Jump to
12) ST_SYSTEM_OK=ON
DIAG I / O 12) ST_SYSTEM_OK = ON
PREV NEXT JUMP ON PRNT EXIT
Exit to return
to the
DIAG menu
Press NEXT & PREV to
move between signal
types.
Pressing the PRNT button will send a formatted
printout to the serial port and can be captured
with a computer or other output device.
Toggle ON/(OFF) button to
change status.
Figure 5-17: DIAG – Signal I/O Menu
06807C DCN6650