9. CALIBRATION PROCEDURES
This section describes the calibration procedures for the T100. All of the methods
described in this section can be initiated and controlled through the COM ports.
IMPORTANT
IMPACT ON READINGS OR DATA
If you are using the T100 for US-EPA controlled monitoring, refer to
Section 10 for information on the EPA calibration protocol.
9.1. CALIBRATION PREPARATIONS
The calibration procedures in this section assume that the analog range and units of
measure, range mode, and reporting range have already been selected for the analyzer. If
this has not been done, please do so before continuing (refer to Sections 3.4.4.1 and 5 for
instructions).
IMPORTANT
IMPACT ON READINGS OR DATA
It is recommended that the LAMP CAL routine (refer to Section 5.9.6) be
performed prior to all calibration operations. This will allow the
instrument to account for minor changes due to aging of the UV lamp.
9.1.1. REQUIRED EQUIPMENT, SUPPLIES, AND EXPENDABLES
Calibration of the T100 analyzer requires a certain amount of equipment and supplies.
These include, but are not limited to, the following:
Zero-air source
Sulfur dioxide span gas source
Gas lines - all gas line materials should be Teflon-type or glass.
A recording device such as a strip-chart recorder and/or data logger (optional).
Traceability Standards