APPENDIX A-3: Warnings and Test Functions
Table A-2: Warning Messages
NAME MESSAGE TEXT DESCRIPTION
Warnings
WSYSRES SYSTEM RESET Instrument was power-cycled or the CPU was reset.
WDATAINIT DATA INITIALIZED Data storage was erased.
WCONFIGINIT CONFIG INITIALIZED
Configuration storage was reset to factory configuration or
erased.
WSO2ALARM1
7
SO2 ALARM 1 WARN SO
2
concentration alarm limit #1 exceeded
WSO2ALARM2
7
SO2 ALARM 2 WARN SO
2
concentration alarm limit #2 exceeded
WO2ALARM1
10+7
O2 ALARM 1 WARN O
2
concentration alarm limit #1 exceeded
WO2ALARM2
10+7
O2 ALARM 2 WARN O
2
concentration alarm limit #2 exceeded
WCO2ALARM1
11+7
CO2 ALARM 1 WARN CO
2
concentration alarm limit #1 exceeded
WCO2ALARM2
11+7
CO2 ALARM 2 WARN CO
2
concentration alarm limit #2 exceeded
WPMT PMT DET WARNING
PMT detector outside of warning limits specified by
DETECTOR_LIMIT variable.
WUVLAMP UV LAMP WARNING
UV lamp reading outside of warning limits specified by
DETECTOR_LIMIT variable.
WSAMPFLOW SAMPLE FLOW WARN
Sample flow outside of warning limits specified by
SAMP_FLOW_SET variable.
WSAMPPRESS SAMPLE PRESS WARN
Sample pressure outside of warning limits specified by
SAMP_PRESS_SET variable.
WVACPRESS
5
VACUUM PRESS WARN
Vacuum pressure outside of warning limits specified by
VAC_PRESS_SET variable.
WBOXTEMP BOX TEMP WARNING
Chassis temperature outside of warning limits specified by
BOX_SET variable.
WRCELLTEMP RCELL TEMP WARNING
Reaction cell temperature outside of warning limits specified by
RCELL_SET variable.
WO2CELLTEMP
10
O2 CELL TEMP WARN
O
2
sensor cell temperature outside of warning limits specified by
O2_CELL_SET variable.
WIZSTEMP IZS TEMP WARNING
IZS temperature outside of warning limits specified by IZS_SET
variable.
WPMTTEMP PMT TEMP WARNING
PMT temperature outside of warning limits specified by
PMT_SET variable.
WDARKCAL DARK CAL WARNING Dark offset above limit specified by DARK_LIMIT variable.
WHVPS HVPS WARNING
High voltage power supply output outside of warning limits
specified by HVPS_SET variable.
WDYNZERO CANNOT DYN ZERO
Contact closure zero calibration failed while DYN_ZERO was set
to ON.
WDYNSPAN CANNOT DYN SPAN
Contact closure span calibration failed while DYN_SPAN was set
to ON.
WREARBOARD REAR BOARD NOT DET Rear board was not detected during power up.
WRELAYBOARD RELAY BOARD WARN Firmware is unable to communicate with the relay board.