Data Acquisition System (DAS) and APICOM Teledyne API - T100 UV Fluorescence SO2 Analyzer
166
To define the REPORT PERIOD, follow the instruction shown in Figure 7-3, then
press:
Edit Data Channel Menu
SETUP X.X 0) CONC: ATIMER, 2, 4032, R
PREV NEXT INS DEL EDIT PRNT EXIT
Exits to the main
Data Acquisition
menu.
SETUP X.X NAME:CONC
<SET SET> EDIT PRINT EXIT
Press SET> until you reach REPORT PERIOD
SETUP X.X REPORT PERIODD:DAYS:0
0 0 0 ENTR EXIT
ENTR accepts the new string and
returns to the previous menu.
EXIT ignores the new string and
returns to the previous menu.
Press buttons to set amount of
time between reports, in hours
(HH) and/or minutes (MM)
(max: 23:59). 01:00 sets a
report to be made every hour.
Set the number of days
between reports (0-365).
IIf at any time an invalidl entry is selected (e.g., days > 366)
the ENTR button will disappear from the display.
SETUP X.X REPORT PERIODD:TIME:01:00
0 1 0 0 ENTR EXIT
SETUP X.X REPORT PERIOD:000:00:05
<SET SET> EDIT PRINT EXIT
Use PREV and NEXT to
scroll to the data channel
to be edited.
Figure 7-8: DAS – Define the Report Period
7.2.6. NUMBER OF RECORDS
Although the DAS is capable of capturing several months worth of data,, the actual
number of records is also limited by the total number of parameters and channels and
other settings in the DAS configuration. Every additional data channel, parameter,
number of samples setting etc. will reduce the maximum amount of data points
somewhat. In general, however, the maximum data capacity is divided amongst all
channels (max: 20) and parameters (max: 50 per channel).
The DAS will check the amount of available data space and prevent the user from
specifying too many records at any given point. If, for example, the DAS memory space
can accommodate 375 more data records, the ENTR button will disappear when trying
to specify more than that number of records. This check for memory space may also
make an upload of a DAS configuration with APICOM or a Terminal program fail, if
the combined number of records would be exceeded. In this case, it is suggested to
either try from the front panel what the maximum number of records can be or use trial-
and-error in designing the DAS script or calculate the number of records using the DAS
or APICOM manuals. To set the number of records for one channel from the front panel,
press SETUP-DAS-EDIT-ENTR and the following control button sequence.