Teledyne T100 Camera Accessories User Manual


 
SETUP Menu Teledyne API - T100 UV Fluorescence SO2 Analyzer
126
5.9.5. ELECTRICAL TEST
The electrical test function creates a current, which substitutes the PMT signal, and
feeds it into the preamplifier board. This signal is generated by circuitry on the pre-
amplifier board itself and tests the filtering and amplification functions of that assembly
along with the A/D converter on the motherboard. It does not test the PMT itself. The
electrical test should produce a PMT signal of about 2000 ±1000 mV.
Access the Electrical Test from the DIAG Menu (refer to Figure 5-16), then press:
DIAG SIGNAL I / O
PREV NEXT JUMP ENTR EXIT
DIAG ELEC RANGE = 500.0 PPB SO2 =XXX.X
<TST TST> EXIT
Press NEXT until…
DIAG ELECTRICAL TEST
PREV NEXT ENTR EXIT
Press TST until…
DIAG ELEC PMT = 1732 MV
SO2=X.XXX
<TST TST> EXIT
While the electrical test is
activated, PMT should equal:
2000 mV ± 1000 mV
Figure 5-31: DIAG – Electrical Test
06807C DCN6650