Troubleshooting & Service Teledyne API - T100 UV Fluorescence SO2 Analyzer
260
7. It may be necessary to straighten the pins somewhat to fit them into the socket.
Press the DOM all the way in and reinsert the offset clip.
8. Close the rear panel and turn on power to the machine.
9. If the replacement DOM carries a firmware revision, re-enter all of the setup
information.
12.7.2. SENSOR MODULE REPAIR & CLEANING
CAUTION - GENERAL SAFETY HAZARD
Do not look at the UV lamp while the unit is operating. UV light can cause
eye damage. Always use safety glasses made from UV blocking material
when working with the UV Lamp Assembly. (Generic plastic glasses are
not adequate).
Figure 12-6: Sensor Module Wiring and Pneumatic Fittings
06807C DCN6650