Teledyne T100 Camera Accessories User Manual


 
Teledyne API - T100 UV Fluorescence SO2 Analyzer Calibration Procedures
203
9.10. CALIBRATION OF OPTIONAL SENSORS
This section presents calibration procedures for the O
2
sensor option and for the CO
2
sensor option.
9.10.1. O
2
SENSOR CALIBRATION
Calibration begins with connecting the zero and span gases, then setting the
concentration values.
9.10.1.1. O
2
CALIBRATION SETUP
Bottled gases are connected as follows:
Figure 9-13: O
2
Sensor Calibration Set Up
O
2
SENSOR ZERO GAS: Teledyne API recommends using pure N
2
when calibrating
the zero point of your O
2
sensor option.
O
2
SENSOR SPAN GAS: Teledyne API recommends using 20.9% O
2
in N
2
when
calibration the span point of your O
2
sensor.
9.10.1.2. SET O
2
SPAN GAS CONCENTRATION
Set the expected O
2
span gas concentration.
This should be equal to the percent concentration of the O
2
span gas of the selected
reporting range (default factory setting = 20.9%; the approximate O
2
content of ambient
air).