Teledyne T100 Camera Accessories User Manual


 
89
5. SETUP MENU
The SETUP menu is used to set instrument parameters for performing configuration,
calibration, reporting and diagnostics operations according to user needs.
5.1. SETUP – CFG: CONFIGURATION INFORMATION
Pressing the CFG button displays the instrument configuration information. This display
lists the analyzer model, serial number, firmware revision, software library revision,
CPU type and other information. Use this information to identify the software and
hardware when contacting Technical Support. Special instrument or software features or
installed options may also be listed here.
SAMPLE RANGE = 500.000 PPB SO2 =XXX.X
< TST TST > CAL SETUP
SAMPLE ENTER SETUP PASS : 818
8 1 8 ENTR EXIT
SAMPLE PRIMARY SETUP MENU
CFG DAS RNGE PASS CLK MORE EXIT
SAMPLE T100 SO2 ANALYZER
NEXT PREV EXIT
Press EXIT at
any time to
return to
SETUP menu
Press EXIT
at
any time to
return to the
SAMPLE dis
p
la
y
Press NEXT of PREV to move back
and forth through the following list
of Configuration information:
MODEL NAME
SERIAL NUMBER
SOFTWARE REVISION
LIBRARY REVISION
iCHIP SOFTWARE REVISION
1
HESSEN PROTOCOL REVISION
1
ACTIVE SPECIAL SOFTWARE
OPTIONS
1
CPU TYPE
DATE FACTORY CONFIGURATION
SAVED
1
Only appears if relevant option of Feature is active.
Figure 5-1: SETUP – Configuration Information
5.2. SETUP – ACAL: AUTOMATIC CALIBRATION OPTION
The menu button for this option appears only when the instrument has the zero span
and/or IZS options. See Section 9.8 for details.
06807C DCN6650