Teledyne API - T100 UV Fluorescence SO2 Analyzer SETUP Menu
111
5.9.2. ANALOG OUTPUT STEP TEST
Analog Output is used as a step test to check the accuracy and proper operation of the
analog outputs. The test forces all four analog output channels to produce signals
ranging from 0% to 100% of the full scale range in 20% increments. This test is useful
to verify the operation of the data logging/recording devices attached to the analyzer.
Access the Analog Output Step Test from the DIAG Menu (refer to Figure 5-16), then
press:
SAMPLE RANGE = 500.000 PPB SO2 =XXX.X
< TST TST > CAL SETUP
SAMPLE ENTER SETUP PASS : 818
8 1 8 ENTR EXIT
SETUP X.X
PRIMARY SETUP MENU
CFG DAS RNGE PASS CLK MORE EXIT
DIAG SIGNAL I / O
NEXT ENTR EXIT
Exit-Exit
returns to the
DIAG menu
SETUP X.X SECONDARY SETUP MENU
COMM VARS DIAG EXIT
DIAG
A
NALOG OUTPUT
PREV NEXT ENTR EXIT
DIAG AOUT ANALOG OUTPUT
0% EXIT
Performs
analog output
step test.
0% - 100%
DIAG AOUT ANALOG OUTPUT
[0%] EXIT
Pressing the “0%” button while performing the test will
pause the test at that level. Brackets will appear around
the value: example: [20%] Pressing the same key again
will resume the test.
Figure 5-18: DIAG – Analog Output Menu
06807C DCN6650