Principles of Operation Teledyne API - T100 UV Fluorescence SO2 Analyzer
300
13.5.2. SENSOR MODULE
Electronically, the T100 sensor module is a group of components that: create the UV
light that initiates the fluorescence reaction between SO
2
and O
3
; sense the intensity of
that fluorescence; generate various
electronic signals needed by the analyzer to
determine the SO
2
concentration of the sample gas (refer to Section 13.1.1), and sense
and control key environmental conditions such as the temperature of the sample gas and
the PMT.
Figure 13-15: T100 Sensor Module
These components are divided into two significant subassemblies: the sample chamber
and the PMT assembly.
Figure 13-16 shows an exploded view of the sample chamber assembly
Figure 13-17shows an exploded view of the PMT Assembly
06807C DCN6650