Teledyne T100 Camera Accessories User Manual


 
APPENDIX A - Version Specific Software Documentation
Rev 1.0.3 (T-Series)/G6 (E-Series)
Teledyne API - Models T100, 100E Series (05036F DCN6650)
A-20 05036 Rev D
SETUP VARIABLE
NUMERIC
UNITS
DEFAULT
VALUE
VALUE RANGE DESCRIPTION
FACTORY_OPT BitFlag 0 0–65535
Factory option flags. Add values to
combine flags.
1 = enable dilution factor
2 = zero/span valves installed
4 = IZS installed (implies zero/span
valves installed)
8 = low span valve installed
16 = display units in concentration field
32 = enable software-controlled
maintenance mode
64 = enable lamp power analog output
128 = enable switch-controlled
maintenance mode
256 = compute only offset during zero
calibration
1024 = enable high flow rate sensor
2048 = enable Internet option
4096 = enable pre-amplified UV lamp
monitoring
8192 = enable non-zero offset
calibration
16384 = enable pressurized zero
calibration
32768 = enable pressurized span
calibration
1
T100/M100E.
2
M100ES.
3
T100H/M100EH.
4
Background concentration compensation option (6400E/6400EH).
6
Engineering firmware only.
7
iChip option.
8
Must power-cycle instrument for these options to fully take effect.
9
Low span option.
10
O
2
option.
11
CO
2
option.
12
T100U/M100EU.
13
Concentration alarm option.
17
TAI protocol
18
Aging Compensation option.
99
Obsolete.
06807C DCN6650