10.6. SPECIAL CALIBRATION REQUIREMENTS FOR DUAL
RANGE OR AUTO RANGE
If Dual Range or Auto Range is selected, then calibration for Range1 and Range2
should be performed separately.
For zero and span point calibration, follow the procedure described in Section 9.2.
Repeat the procedure for both the HIGH and LOW Ranges.
10.7. REFERENCES
1. Environmental Protection Agency, Title 40, Code of Federal Regulations, Part 50,
Appendix A, Section 10.3.
2. Quality Assurance Handbook for Air Pollution Measurement Systems - Volume II,
Ambient Air Specific Methods, EPA-600/4-77-027a, 1977.
3. Catalog of NBS Standard Reference Materials. NBS Special Publication 260, 1975-
76 Edition. U.S. Department of Commerce, NBS. Washington, D.C. June 1975. (Tel:
301-975-6776 for ordering the catalog).
4. Quality Assurance Handbook for Air Pollution Measurement Systems - Volume I,
Principles. EPA-600/9-76-005. March 1976.
.