Teledyne API - Models T100, 100E Series (05036F DCN6650) APPENDIX A - Version Specific Software Documentation
Rev 1.0.3 (T-Series)/G6 (E-Series)
A-39
MODBUS
Register Address
(dec., 0-based)
Description Units
MODBUS Discrete Input Registers
(single-bit; read-only)
0 PMT detector warning
1 UV detector warning
2 Dark calibration warning
3 Box temperature warning
4 PMT temperature warning
5 Reaction cell temperature warning
6 Sample pressure warning
7 HVPS warning
8 System reset warning
9 Rear board communication warning
10 Relay board communication warning
11 Front panel communication warning
12 Analog calibration warning
13 Dynamic zero warning
14 Dynamic span warning
15 Invalid concentration
16 In zero calibration mode
17 In span calibration mode
18 In multi-point calibration mode
19 System is OK (same meaning as SYSTEM_OK I/O signal)
20 Sample flow warning
21
1
IZS temperature warning
22
2
In low span calibration mode
23
2
Vacuum pressure warning
24
3
SO
2
concentration alarm limit #1 exceeded
25
3
SO
2
concentration alarm limit #2 exceeded
26 In Hessen manual mode
100
10
In O
2
calibration mode
101
10
O
2
cell temperature warning
102
10+3
O
2
concentration alarm limit #1 exceeded
103
10+3
O
2
concentration alarm limit #2 exceeded
200
11
In CO
2
calibration mode
201
11+3
CO
2
concentration alarm limit #1 exceeded
202
11+3
CO
2
concentration alarm limit #2 exceeded