Teledyne T100 Camera Accessories User Manual


 
Calibration Procedures Teledyne API - T100 UV Fluorescence SO2 Analyzer
192
9.4. MANUAL CALIBRATION WITH ZERO/SPAN VALVES
Zero and Span calibrations using the Zero/Span Valve option are similar to that
described in Section 7.2, except that:
Zero air and span gas are supplied to the analyzer through the zero gas and span gas
inlets rather than through the sample inlet.
The zero and cal operations are initiated directly and independently with dedicated
buttons (CALZ and CALS)
STEP ONE: Connect the sources of zero air and span gas to the respective ports on the
rear panel (refer to Figure 3-1) as shown below.
External Zero
Air Scrubber
Filter
Source of
SAMPLE Gas
VENT if input is pressurized
MODEL 700
Gas Dilution Calibrator
(with O
3
generator option)
VENT
SAMPLE
ZERO AIR
EXHAUST
SPAN
1
Calibrated
SO
2
Gas
(At high
concentration)
Needle valve
to control flow
VENT
MODEL 701
Zero Air
Generator
Chassis
Figure 9-5: Setup for Manual Calibration with Z/S Valve Option Installed (Step 1)
06807C DCN6650