5.4.3.1. SINGLE RANGE MODE (SNGL)
The default range mode for the analyzer is single range, in which all analog
concentration outputs are set to the same reporting range. This reporting range can be set
to any value between 0.1 ppb and 20,000 ppb.
While the two outputs always have the same reporting range, the span and scaling of
their electronic signals may also be configured for different differently (e.g., A1 = 0-10
V; A2 = 0-0.1 V).
To select SNGLE range mode and to set the upper limit of the range, press:
SAMPLE ENTER SETUP PASS : 818
8 1 8 ENTR EXIT
SETUP X.X
PRIMARY SETUP MENU
CFG DAS RNGE PASS CLK MORE EXIT
SETUP X.X RANGE CONTROL MENU
MODE
SET UNIT EXIT
SETUP X.X RANGE MODE: SNGL
SNGL
IND AUTO ENTR EXIT
EXIT x 2 returns
to the main
SAMPLE display
SETUP X.X RANGE MODE: SNGL
SNGL
IND AUTO ENTR EXIT
SETUP X.X RANGE CONTROL MENU
MODE
SET UNIT EXIT
SETUP X.X RANGE: 500.0 Conc
0 0 5 0 0 .0 ENTR EXIT
SETUP X.X RANGE CONTROL MENU
MODE
SET UNIT EXIT
SAMPLE RANGE = 500.000 PPB SO2 =XXX.X
< TST TST > CAL SETUP
Figure 5-4: SETUP RNGE – Single Range Mode
06807C DCN6650