Teledyne T100 Camera Accessories User Manual


 
Teledyne API - T100 UV Fluorescence SO2 Analyzer SETUP Menu
117
To select manual output calibration for a particular channel, access the Analog I/O
Configuration from the DIAG Menu (refer to Figure 5-16), then press:
DIAG AIO CONC_OUT_2 AUTO CAL: ON
ON ENTR EXIT
Exit to return to
the main
sample display
DIAG ANALOG I / O CONFIGURATION
PREV NEXT ENTR EXIT
DIAG AIO AOUTS CALIBRATED: NO
< SET SET> CAL EXIT
DIAG AIO CONC_OUT_2 RANGE: 5V
SET> EDIT EXIT
DIAG AIO CONC_OUT_2:5V, CAL
< SET SET> EDIT EXIT
Press SET> to select the analog output channel to
be configured. Then press EDIT to continue
ENTR accepts the new setting
and returns to the previous menu.
EXIT ignores the new setting and
returns to the previous menu.
DIAG AIO CONC_OUT_2 REC OFS: 0 mV
< SET SET> EDIT EXIT
DIAG AIO CONC_OUT_2 AUTO CAL: ON
< SET SET> EDIT EXIT
DIAG AIO CONC_OUT_2 AUTO CAL: OFF
OFF ENTR EXIT
Toggle to OFF to enable manual
cal adjustments to the selected
out
p
ut channel onl
y
.
Figure 5-22: DIAG – Analog Output – Auto Cal or Manual Cal Selection for Channels
Now the analog output channels should either be automatically calibrated or they should
be set to manual calibration, which is described next.
06807C DCN6650