Teledyne API - T100 UV Fluorescence SO2 Analyzer Troubleshooting & Service
237
The analyzer also issues a message via the serial port(s).
To view or clear a warning message press:
SAMPLE RANGE = 500.0 PPB SO2 =XXX.X
TEST CAL MSG CLR SETUP
If warning messages reappear,
the cause needs to be found. Do
not repeatedly clear warnings
without corrective action.
Press CLR to clear the current
warning message.
If more than one warning is
active, the next message will
take its place.
Once the last warning has been
cleared, the analyzer returns to
SAMPLE Mode.
SAMPLE RANGE = 500.0 PPB SO2 =XXX.X
< TST TST > CAL MSG CLR SETUP
SAMPLE SYSTEM RESET SO2= X.XXX
< TST TST > CAL MSG CLR SETUP
In WARNING mode, <TST TST>
buttons replaced with TEST
button. Pressing TEST switches to
SAMPLE mode and hides warning
messages until new warning(s)
are activated.
MSG indicates that one or more
warning message are active but
hidden. Pressing MSG cycles
through warnings
In SAMPLE mode, all warning
messages are hidden, but MSG
button appears
Figure 12-1: Viewing and Clearing Warning Messages
06807C DCN6650