Getting Started Teledyne API - T100 UV Fluorescence SO2 Analyzer
70
Table 3-14: Possible Startup Warning Messages – T100 Analyzers with Options
Message Meaning
O2 CELL TEMP WARN
1
O
2
sensor cell temperature outside of warning limits specified by
O2_CELL_SET variable.
IZS TEMP WARNING
2
On units with IZS options installed: The permeation tube temperature is outside
of specified limits.
O2 ALARM 1 WARN
1, 4
O
2
Alarm limit #1 has been triggered.
4
O2 ALARM 2 WARN
1, 4
O
2
Alarm limit #2 has been triggered.
4
CO2 ALARM 1 WARN
3, 4
CO
2
Alarm limit #1 has been triggered.
4
CO2 ALARM 2 WARN
3, 4
`
CO
2
Alarm limit #2 has been triggered.
4
SO2 ALARM1 WARN
4
SO
2
Alarm limit #1 has been triggered.
4
SO2 ALARM2 WARN
4
SO
2
Alarm limit #2 has been triggered.
4
1
Only appears when the optional O
2
sensor is installed.
2
Only appears when the optional internal zero span (IZS) option is installed.
3
Only appears when the optional CO
2
sensor is installed.
4
Only Appears when the optional gas concentration alarms are installed
3.4.3. FUNCTIONAL CHECKS
After the analyzer’s components have warmed up for at least 60 minutes, verify that the
software properly supports any hardware options that were installed.
For information on navigating through the analyzer’s software menus, refer to the menu
trees described in Appendix A.1.
Check to ensure that the analyzer is functioning within allowable operating parameters.
Appendix C includes a list of test functions viewable from the analyzer’s front panel
as well as their expected values.
These functions are also useful tools for diagnosing performance problems with
your analyzer (refer to Section 12.1.2).
The enclosed
Final Test
and Validation Data Sheet (P/N 04551) lists these values
before the instrument left the factory.
06807C DCN6650