Teledyne T100 Camera Accessories User Manual


 
Teledyne API - Models T100, 100E Series (05036F DCN6650) APPENDIX A - Version Specific Software Documentation
Rev 1.0.3 (T-Series)/G6 (E-Series)
05036 Rev D A-27
SIGNAL NAME BIT OR CHANNEL
NUMBER
DESCRIPTION
B status outputs, U27, J1018, pins 18 = bits 07, default I/O address 324 hex
ST_LAMP_ALARM 0 0 = lamp intensity low
1 = lamp intensity OK
ST_DARK_CAL_ALARM 1 0 = dark cal. warning
1 = dark cal. OK
ST_FLOW_ALARM 2 0 = any flow alarm
1 = all flows OK
ST_PRESS_ALARM 3 0 = any pressure alarm
1 = all pressures OK
ST_TEMP_ALARM 4 0 = any temperature alarm
1 = all temperatures OK
ST_HVPS_ALARM 5 0 = HVPS alarm
1 = HVPS OK
ST_CO2_CAL
11
6 0 = in CO
2
calibration
1 = not in CO
2
calibration
ST_O2_CAL
10
7 0 = in O
2
calibration
1 = not in O
2
calibration
Front panel I
2
C keyboard, default I
2
C address 4E hex
MAINT_MODE 5 (input) 0 = maintenance mode
1 = normal mode
LANG2_SELECT 6 (input) 0 = select second language
1 = select first language (English)
SAMPLE_LED 8 (output) 0 = sample LED on
1 = off
CAL_LED 9 (output) 0 = cal. LED on
1 = off
FAULT_LED 10 (output) 0 = fault LED on
1 = off
AUDIBLE_BEEPER 14 (output) 0 = beeper on (for diagnostic testing only)
1 = off