Appendix C
Warranty/Repair
Questionnaire
T100, M100E
(04796F DCN6611)
CUSTOMER: ____
_____________________________ PHONE: __________________________________________________
TELEDYNE API CUSTOMER SERVICE
Email: SDA_techsupport@teledyne.com
PHONE: (858) 657-9800 TOLL FREE: (800) 324-5190 FAX: (858) 657-9816
C-1
CONTACT NAME: ____________________________ FAX NO. __________________________________________________
SITE ADDRESS: _________________________________________________________________________________________
MODEL SERIAL NO.: ______________________ FIRMWARE REVISION: ______________________________________
1. ARE THERE ANY FAILURE MESSAGES? _______________________________________________________________
________________________________________________________________________________________________________
________________________________________________________________________________________________________
2. PLEASE COMPLETE THE FOLLOWING TABLE: (NOTE: DEPENDING ON OPTIONS INSTALLED, NOT ALL TEST
PARAMETERS BELOW WILL BE AVAILABLE IN YOUR INSTRUMENT)
*IF OPTION IS INSTALLED
Parameter Recorded Value Acceptable Value
RANGE
PPB/PPM 50 PPB to 20 PPM
STABIL
PPB
1 PPB WITH ZERO AIR
SAMP PRESS
IN-HG-A ~ 2” < AMBIENT
SAMPLE FLOW
cm
3
/MIN 650 ± 10%
PMT SIGNAL WITH
ZERO AIR
mV -20 TO 150 mV
PMT SIGNAL AT
SPAN GAS CONC
mV
PPB/PPM
0-5000 mV
0-20000 PPB
NORM PMT AT
SPAN GAS CONC
mV
PPB/PPM
0-5000 mV
0-20000 PPB
UV LAMP
mV 1000 TO 4800 mV
LAMP RATIO
mV 30 TO 120%
STR. LGT
PPB
≤ 100 PPB/ ZERO AIR
DARK PMT
mV -50 TO 200 mV
DARK LAMP
mV -50 TO 200 mV
SLOPE
1.0 ± 0.3
OFFSET
mV < 250 mV
HVPS
V
≈ 400 – 900
RCELL TEMP
ºC 50ºC ± 1
BOX TEMP
ºC AMBIENT + ~ 5
PMT TEMP
ºC 7ºC ± 2º CONSTANT
IZS TEMP*
ºC 50ºC ± 1
ETEST
mV 2000 mV ± 1000
OTEST
mV 2000 mV ± 1000
Values are in the Signal I/O
REF_4096_MV
mV 4096mv±2mv and Must be Stable
REF_GND
mV 0± 0.5 and Must be Stable
3. WHAT IS THE SAMPLE FLOW & SAMPLE PRESSURE W/SAMPLE INLET ON REAR OF MACHINE CAPPED?
SAMPLE FLOW -
CC SAMPLE PRESS - IN-HG-A
06807C DCN6650