Teledyne T100 Camera Accessories User Manual


 
Teledyne API - T100 UV Fluorescence SO2 Analyzer Calibration Procedures
197
STEP TWO: Perform the zero/span check.
Wait until STABIL
falls below 0.5
ppb. This may
take several
minutes.
The value of STABIL
may jump
significantly. Wait
until STABIL falls
below 0.5 ppb. This
may take several
minutes.
SAMPLE STABIL=XXX.X PPB SO2 =XXX.X
< TST TST > CAL CALZ CALS SETUP
ZERO CAL M STABIL=XXX.X PPB SO2 =XXX.X
< TST TST > ZERO CONC EXIT
SAMPLE STABIL=XXX.X PPB SO2 =XXX.X
< TST TST > CAL CALZ CALS SETUP
EXIT returns to the main
SAMPLE dis
p
la
y
ACTION:
Record the
SO
2
readings
presented in the
upper right corner of
the display.
ACTION:
Record the
SO
2
readings
presented in the
upper right corner of
the display.
Scroll to the STABIL test
function.
SAMPLE RANGE = 500.000 PPB SO2 =XXX.X
< TST TST > CAL CALZ CALS SETUP
SPAN CAL M STABIL=XXX.X PPB SO2 =XXX.X
< TST TST > ZERO SPAN CONC EXIT
Figure 9-10: Setup for Manual Calibration Check with Z/S Valve or IZS Option (Step 2)
06807C DCN6650