SETUP Menu Teledyne API - T100 UV Fluorescence SO2 Analyzer
116
To automatically calibrate a single analog channel from the DIAG Menu (refer to Figure
5-16), press:
DIAG AIO CONC_OUT_2 CALIBRATED: NO
<SET CAL EXIT
DIAG AIO AUTO CALIBRATING CONC_OUT_2
DIAG AIO CONC_OUT_2 CALIBRATED: YES
<SET CAL EXIT
DIAG ANALOG I / O CONFIGURATION
PREV NEXT ENTR EXIT
DIAG AIO AOUTS CALIBRATED: NO
< SET> CAL EXIT
DIAG AIO CONC_OUT_2 RANGE: 5V
SET> EDIT EXIT
DIAG AIO CONC_OUT_2:5V, CAL
< SET SET> EDIT EXIT
DIAG AIO CONC_OUT_2 REC OFS: 0 mV
< SET SET> EDIT EXIT
DIAG AIO CONC_OUT_2 AUTO CAL: ON
< SET SET> EDIT EXIT
EXIT to Return
to the main
Sample Display
Press SET> to select the
Analog Output channel to
be configured. Then Press
EDIT to continue
Figure 5-21: DIAG – Analog Output Calibration Mode – Single Analog Channel