Teledyne API - T100 UV Fluorescence SO2 Analyzer Specifications, Approvals & Compliance
31
2.2. EPA EQUIVALENCY DESIGNATION
The T100 Analyzer is designated as Reference Method Number EQSA-0495-100 as per
40 CFR Part 53 when operated under the following conditions:
Range: Any range from 50 parts per billion (ppb) to 10 parts per million (ppm)
Ambient temperature range of 5
o
C to 40
o
C
Line voltage range of 100-120 VAC or 220-240 VAC, at 50 or 60 Hz
Sample filter: Equipped with PTFE filter element in the internal filter assembly
Sample flow of 650 +/- 65 cm
3
/min
Vacuum pump (internal) capable of 14"Hg Absolute pressure @ 1 slpm or better
Software settings:
Dynamic span OFF
Dynamic zero OFF
Dilution factor OFF
AutoCal ON or OFF
Dual range ON or OFF
Auto-range ON or OFF
Temp/Pressure compensation ON
Under this designation, the analyzer may be operated with or without the following
optional equipment:
Rack mount with chassis slides
Rack mount without slides, ears only
Zero/span valve options.
Internal zero/span (IZS) option with either:
SO
2
permeation tube - 0.4ppm at 0.7 liter per minute; certified/uncertified,
or
SO
2
permeation tube - 0.8 ppm at 0.7 liter per minute; certified/uncertified.
Under the designation, the IZS option cannot be used as the source of
calibration
4-20mA isolated analog outputs
Status outputs
Control inputs
RS-232 output
Ethernet output
Zero air scrubber
4-20mA, isolated output