Teledyne API - T100 UV Fluorescence SO2 Analyzer Troubleshooting & Service
263
To clean the sample chamber:
1. Remove the sensor module as described in Section 12.7.2.1.
2. Remove the sample chamber mounting bracket by unscrewing the four bracket
screws.
Figure 12-8: Sample Chamber Mounting Bracket
3. Unscrew the 4 hexagonal standoffs.
4. Gently remove the chamber cover.
5. Using a lint-free cloth dampened with distilled water, wipe the inside surface of the
chamber and the chamber cover.
6. Dry the chamber surfaces with a 2nd lint-free cloth.
7. Re-assemble the chamber and re-install the sensor module.
12.7.2.3. CLEANING THE PMT LENS AND PMT FILTER
IMPORTANT
IMPACT ON READINGS OR DATA
The sample chamber should only be opened or cleaned on instructions
from the Teledyne API Technical Support Department.
Be careful not to leave thumbprints on the interior of the sample
chamber. The various oils that make up fingerprints fluoresce brightly
under UV light and will significantly affect the accuracy of the
analyzer’s SO
2
measurement).